Oxygen Monitoring — SpO2–SaO2 Discordance

Current · V5 (2025) → C9.i Historical · V4 (2023) → F12.i 1 exam appearance

2023B Q01

Exam question

Explain why the oxygen-haemoglobin saturation value derived by a pulse oximeter (SpO2) could be different from the measured arterial value (SaO2).

CICMWrecks answer

Master answer

Methods of Measuring oxygen saturation

Co-oximetry

Pulse oximetry

SpO2 different from SaO2

PATIENT FACTORS
Low or High SpO2Low SpO2Normal or High SpO2
Met-Hb
Sulph-Hb
Poor perfusion of finger
Movement artifact
Venous pulsations
Fingernail polish
Intravenous pigmented dyes
Haemoglobinopathy
Anaemia with co-existing hypoxia
Carbon Monoxide poisoning
EQUIPMENT FACTORS
– Ambient light interference
– Poorly fitting probe
– Assay calibrated using healthy volunteers only down to SpO2 80%. Unknown significance if tested SpO2 less than 80%
PHYSIOLOGICAL FACTORS
– Due to O2 dissociation curve, insensitive to changes above PaO2 80mmHg
– Does not measure tissue oxygenation