Past Papers · SAQ

Carbon Dioxide Monitoring — Capnography Principles

Current · V5 (2025) → C9.ii Historical · V4 (2023) → F12.ii 5 exam appearances

2026A Q11

Exam question

a) Outline the scientific principles that apply to the measurement of end-tidal carbon dioxide using capnography. Include the techniques of sampling in your answer (30% of marks).
b) Describe a normal capnograph waveform and its features (20% of marks). A diagram may assist you with your answer.
c) Outline the ventilation and perfusion information that can be derived from the capnograph waveform (50% of marks).

CICMWrecks answer

Master answer

a) Scientific principles and sampling (30%)

Measurement principle

Sampling techniques

Feature Mainstream / in-line Sidestream
Method Infrared sensor sits directly in the breathing circuit near the airway. A small continuous gas sample is aspirated through tubing to a remote analyser.
Advantages Very rapid response; minimal transport delay; waveform is less distorted by sampling tubing. Lightweight airway adapter; can be used with non-intubated patients via specialised cannulae; analyser is remote from the airway.
Limitations Adds apparatus weight/dead space; secretions or condensation can contaminate the optical window. Sampling delay and waveform dispersion; tubing may kink, leak or block with water/secretions; aspiration removes a small amount of circuit gas.

Important measurement errors

b) Normal capnogram waveform and features (20%)

Normal time capnogram showing inspiratory baseline, expiratory phases I to III, end-tidal carbon dioxide, alpha angle and beta angle.
CICMWrecks redraw (2026) · Recreated from legacy figure

Time capnogram: CO2 partial pressure/concentration on the y-axis versus time on the x-axis.

Phase Normal feature Physiological meaning
Phase 0 — inspiration Rapid downstroke to a baseline near zero. CO2-free inspired gas replaces alveolar gas.
Phase I Inspiratory/early expiratory baseline near zero. Exhalation of apparatus and anatomical dead-space gas containing essentially no CO2.
Phase II Rapid expiratory upstroke. Mixing of dead-space gas with progressively increasing alveolar gas.
Phase III Alveolar plateau with a small positive slope. Predominantly alveolar gas; the slope reflects sequential emptying of alveoli with different time constants and V/Q ratios.

c) Ventilation and perfusion information derived from the waveform (50%)

Ventilation

Capnographic finding What it tells you Why
Persistent exhaled CO2 Supports tracheal rather than oesophageal intubation. Pulmonary ventilation produces repeated alveolar CO2 waveforms; an oesophageal tube does not sustain them.
Respiratory rate Breath frequency, apnoea and sudden circuit disconnection can be recognised immediately. Each ventilatory cycle generates one capnogram.
ETCO2 trend Reflects adequacy of alveolar ventilation when CO2 production and perfusion are reasonably stable. PaCO2 is inversely related to alveolar ventilation; ETCO2 usually tracks PaCO2.
Rising ETCO2 May indicate hypoventilation or increased CO2 production. Less alveolar ventilation removes less CO2.
Falling ETCO2 May indicate hyperventilation, but must also prompt consideration of reduced pulmonary perfusion. ETCO2 depends on both ventilation and delivery of CO2 to the lung.
Steep phase III / increased alpha angle Airflow obstruction or heterogeneous alveolar emptying. Different regional time constants cause sequential emptying of alveoli with differing CO2 concentrations.
Inspiratory baseline above zero / increased beta angle CO2 rebreathing. Inspired gas contains residual CO2 rather than returning to a zero baseline.

Perfusion

Integrated V/Q information

Exam focus

Quick reference

Summary

FeatureHigh-yield point
MeasurementInfrared absorption at ~4.26 μm using the Beer–Lambert relationship
SamplingMainstream = in-line rapid response; sidestream = remote analyser with transport delay / sampling-line errors
Normal waveformPhase 0 inspiration; I dead space; II expiratory upstroke; III alveolar plateau; ETCO2 at end of III
VentilationETCO2 trend, respiratory rate, tube position, rebreathing, obstruction and heterogeneity of alveolar emptying
PerfusionWith ventilation/metabolism stable, ETCO2 trends with pulmonary blood flow / cardiac output; PaCO2−ETCO2 gradient widens with alveolar dead space

Past papers

Exam appearances

5 appearances
Exam Exact exam wording Candidate success
2026A Q11 a) Outline the scientific principles that apply to the measurement of end-tidal carbon dioxide using capnography. Include the techniques of sampling in your answer (30% of marks). b) Describe a normal capnograph waveform and its features (20% of marks). A diagram may assist you with your answer. c) Outline the ventilation and perfusion information that can be derived from the capnograph waveform (50% of marks). 48.2%
2023A Q03 Outline the principles of measurement of end-tidal CO2 using infrared radiation (25% of Marks). Describe the potential sources of error when using this modality and how they may be mitigated (75% of Marks). 21%
2019B Q10 Describe the principles of capnography, including calibration, sources of error and limitations. 31%
2015B Q09 Describe the principles of measurement of end-tidal CO2, including the sources of error. 19%
2010A Q04 Describe the underlying principles involved in the measurement of end tidal CO2 (by infrared analysis), including sources of error and interference. 10%